New applications in scientific geological work centering around the exact isotopic analysis of Nd and Sr challenges the capabilities of the existing mass spectrometers. The new instrument equipped with 11 detectors for simultaneous measurements of isotope ratios has been designed for the purpose of improving the accuracy of the determination of Nd and Sr isotope ratios to 2 ppm. Following the idea that the accuracy mainly depends on the quality of the detector system immediately leads to a design in which oversized static Faraday-cups are combined with a variable mass dispersion of the analyzer. This will be achieved by using two quadrupole doublets in front of and behind a straight through ion optical element like the well known Wien Filter. The Wien Filter consists basically of a magnetic field B and a superimposed electrostatic field E, arranged perpendicular to each other. Ions of arbitrary mass and charge can move along the straight z axis if their velocity equals v{sub 0} = E{sub 0}/B{sub 0}. Since the velocity of the ions depends on their charge, mass and the potential in the ion source, one concludes, that mass scanning is possible by varying the electrostatic field, keeping the magnetic field constant. Thus,more » rapid scanning is possible. In addition, their is no high mass limitation, contrary to magnetic sector field mass spectrometry.« less