Collision cross section (CCS) values determined in ion mobility-mass spectrometry (IM-MS) are increasingly employed as additional descriptors in metabolomics studies. CCS values must therefore be reproducible and the causes of deviations must be carefully known and controlled. Here, we analyzed lipid standards by trapped ion mobility spectrometry-mass spectrometry (TIMS-MS) to evaluate the effects of solvent and flow rate in flow injection analysis (FIA), as well as electrospray source parameters including nebulizer gas pressure, drying gas flow rate, and temperature, on the ion mobility and CCS values. The stability of ion mobility experiments was studied over 10 h, which established the need for a delay-time of 20 min to stabilize source parameters (mostly pressure and temperature). Modifications of electrospray source parameters induced shifts of ion mobility peaks and even the occurrence of an additional peak in the ion mobility spectra. This behavior could be essentially explained by ion-solvent cluster formation. Changes in source parameters were also found to impact CCS value measurements, resulting in deviations up to 0.8%. However, internal calibration with the Tune Mix calibrant reduced the CCS deviations to 0.1%. Thus, optimization of source parameters is essential to achieve a good desolvation of lipid ions and avoid misinterpretation of peaks in ion mobility spectra due to solvent effects. This work highlights the importance of internal calibration to ensure interoperable CCS values, usable in metabolomics annotation.