The bottleneck of VDD systems is the reliable detection of the small atrial signals by a floating atrial electrode. Fractally iridium coated electrodes offer excellent sensing and pacing performance. In this study, the performance of such a floating atrial lead in P wave sensing and synchronous ventricular stimulation was examined. Atrial pacing was also used as a test of atrial wall contact. A fractally iridium coated VDD lead was implanted in 18 patients. In 15 patients it was interfaced with a VDD pacemaker and in 3 patients with a DDD system depending on the P wave amplitude measured acutely (> or = 2 mV). Simultaneous recordings of the surface ECG and pacemaker telemetry were used to analyze P wave amplitudes and AV synchrony in different body positions, and during normal and deep breathing. Additionally, exercise tests based on daily life activities and 24-hour ECG monitoring were performed to test the pacemaker function. During implantation P wave amplitudes were 1.86 mV +/- 1.08 mV (range 0.5-4.9 mV) and during follow-up (6.6 +/- 5.6 weeks) 0.18-3.8 mV. Holter recordings revealed reliable P wave sensing at a sensitivity setting of 0.5 mV (95.5%). P wave sensing was further improved by a higher atrial sensitivity. AV synchronous pacing > or = 99.9% was achieved in all patients. In 7 patients the atrial electrode could be positioned close to the atrial wall enabling atrial stimulation thresholds at an average of 4.3 volts. This fractally iridium coated VVD lead allowed consistent and reliable P wave sensing at an atrial sensitivity as low as 0.5 mV in selected patients.