The response of a DPbS3200 infrared detector irradiated by a nanosecond pulsed laser and CW laser has been investigated to study laser-induced interference. A laser interference experiment system was constructed to measure the time-varying response signal. A nanosecond pulsed laser and a CW laser of 10 Hz were used, with a 1064 nm wavelength and a millimeter-scale irradiation spot diameter. Firstly, the characteristics of transient interference signals induced by pulsed lasers were analyzed. Then, the characteristics of response signal interference by both CW laser and pulsed laser irradiation were further investigated. The results showed that the pulsed laser only produced transient interference. However, the CW laser led to a significant amplitude reduction of the response signal, which could continuously interfere in the operating time. For transient interferences, the amplitude of the interference signal increased linearly with the laser fluence. The relation between the pulse repetition rate of the incident laser and the operating frequency of the detector determined the numbers of transient interference signals in one response period; for the interference induced by both the CW laser and pulsed laser, CW laser interference played a leading role when CW laser power density increased to 4.1 W/cm2 or more. As the CW laser fluence reached 6.1 W/cm2, the PbS infrared detector was no longer able to detect any signal, which caused temporary blindness. In the end, a probit model was used to determine the interference threshold.