Radiation pneumonitis (RP) is a potential toxic side effect of thoracic radiotherapy. Optimal planning techniques must maintain tumor coverage while limiting dose to normal lung tissue to reduce the risk of patients developing RP. The addition of a noncoplanar arc may be beneficial by increasing treatment angles and providing an ideal dose distribution for tumor coverage while decreasing dose to organs at risk (OAR). The purpose of this research was to compare the effects on the normal bilateral lung tissue receiving 20 Gy, 10 Gy and 5 Gy (V20, V10, V5) and the mean lung dose (MLD) values when medial lung tumors are treated with 3 partial coplanar arcs vs 2 partial coplanar arcs combined with a partial sagittal arc. Researchers hypothesized that a beam arrangement of 2 partial coplanar arcs and 1 partial sagittal arc would reduce V20, V10, V5, and MLD values when compared to a 3 partial coplanar arc plan. In a retrospective study of 5 patients with bulky, medial right lung lesions without nodal involvement, cases were planned with both a noncoplanar and a coplanar arc geometry. Results were evaluated using a two-tailed t-test to determine the statistical significance (p < 0.05) of changes to total lung volume analyzation metrics when a noncoplanar sagittal arc was incorporated compared to the standard lung treatment using only coplanar arcs. Although some patient cases showed minor improvement in the V20, V10, V5, and MLD metrics, the study results were not statistically significant and showed no advantage with the introduction of an anterior sagittal arc over a coplanar beam arrangement.