Continuous noninvasive cardiac output monitoring (NICOM) is a clinically useful tool in the pediatric setting. This study compared the ability of stroke volume variation (SVV) measured by NICOM with that of respiratory variations in the velocity of aortic blood flow (△Vpeak) and central venous pressure (CVP) to predict of fluid responsiveness in mechanically ventilated children after ventricular septal defect repair. The study investigated 26 mechanically ventilated children after the completion of surgery. At 30min after their arrival in an intensive care unit, a colloid solution of 10ml/kg was administrated for volume expansion. Hemodynamic variables, including CVP, stroke volume, and △Vpeak in addition to cardiac output and SVV in NICOM were measured before and 10min after volume expansion. The patients with a stroke volume increase of more than 15% after volume expansion were defined as responders. The 26 patients in the study consisted of 13 responders and 13 nonresponders. Before volume expansion, △Vpeak and SVV were higher in the responders (both p values <0.001). The areas under the receiver operating characteristic curves of △Vpeak, SVV, and CVP were respectively 0.956 (95% CI 0.885-1.00), 0.888 (95% CI 0.764-1.00), and 0.331 (95% CI 0.123-0.540). This study showed that SVV by NICOM and △Vpeak by echocardiography, but not CVP, reliably predicted fluid responsiveness during mechanical ventilation after ventricular septal defect repair in children.