The use of Application Specific Integrated Circuits (ASICs) in nuclear physics instrumentation is drastically increasing, thanks to the possibility of incorporating a large number of acquisition channels in compact devices. In this paper, we describe a SiPM-based application with CAEN Front-End Readout System [1] based on Citiroc 1A chip from Weeroc [2]. Besides the use of this chip for well known single photon spectra and event counting, this paper exploits the possibility to acquire energy spectra directly from scintillators, paired with SiPM, through peak-and-hold readout. In particular, good energy resolutions have been achieved even with slow scintillators, like LYSO, CsI(Tl), and BGO, which have 40 ns, 1000 ns, and 300 ns of light decay time, respectively. These values are of the same order of magnitude of the shaping time of the Citiroc 1A chip (maximum value of 87.5 ns) in the case of LYSO, and higher in the case of CsI(Tl) and BGO. Several measurements have been performed using multiple radioactive γ sources and the resulting energy spectra demonstrate a resolution compatible with that found in literature [4] [5] [6] [7], as well as with an alternative acquisition system based on a digitizer that implements an algorithm of Charge Integration in the FPGA [8].