Abstract Introduction Despite the obvious benefits, the use of implantable cardioverter defibrillators (ICDs) is associated with serious complications, including infections. In particular, device replacement or revision procedures are associated with an increased risk of infection. Aim of the study was to compare the cardiac device infection rate with and without the use of the gentamycin-collagen sponge (GCS) in patients undergoing device replacement or revision procedures of implanted ICDs. Methods A total of 491 consecutive ICD patients from a prospective single-centre ICD-registry who underwent elective device replacement, system upgrade or lead revision between 2017 and 6/2023 were analyzed. From September 2020 onwards, a GCS was inserted into the device pocket (GCS group, n = 256). These patients were compared to the patients implanted/ revised between 2017 and 8/2020 (control group, n = 235). Results The baseline characteristics and the PADIT score were similar in both groups (Table 1). The GCS was well tolerated and there were no complications associated with the GCS. The mean follow-time was 15 + 10 months in the GCS group and 30 + 11 months in the control group. The Kaplan-Meier estimated infection rate after one year was 1.0% in the GCS group and 3.5% in the control group (p = 0.03, Figure 1). Time from revision to infection was 144 + 134 days in the GCS group vs. 362 + 297 days in the control group. Conclusion Adjunctive use of a GCS was associated with a lower incidence of device infections than standard-of-care infection-prevention strategies alone.