Abstract Aim Lag screw cutout is the commonest mechanical failure in dynamic hip screw (DHS) fixation of proximal femoral fractures. Tip-apex distance (TAD), the combined distance from the tip of the lag screw to the apex of the femoral head on AP and lateral views, is the most important predictive factor for cutout. TAD >25mm is associated with a high risk of cutout. The aim was to determine the proportion of DHS fixations performed in our department with ‘acceptable’ TAD (<25mm) and assess the impact of weekly departmental metalwork meetings. Method Retrospective data was collected of all DHS fixations performed in our department in a 6-month period. Intraoperative imaging was reviewed to determine TAD. Posters illustrating TAD recommendations were displayed in theatre, and a weekly departmental metalwork meeting was introduced, reviewing imaging from all surgeries performed. Post-intervention, a further audit cycle over a 6-month period was performed. Results 50 patients underwent DHS fixation during the initial audit cycle. Mean TAD was 22.85mm (range 15mm – 36mm); TAD exceeded 25mm in 13 cases (26%). In the second cycle (n=53), mean TAD reduced to 14.95mm (range 8.5mm – 29.4mm), with 1 case exceeding 25mm (2%). Conclusions The introduction of a weekly departmental metalwork meeting has significantly reduced the mean tip-apex distance of DHS fixations in our department, as well as the proportion of fixations with an unacceptable TAD. This should reduce the risk of lag screw cutout in these patients.
Read full abstract