PurposeTo describe the reliability of ultra-high-resolution computed tomography (UHR-CT) in the measurement of titanium stapes prostheses using manufacturer data as a reference. Materials and methodsThis retrospective study included patients treated by stapedectomy with titanium prostheses who underwent UHR-CT between January 2020 and October 2023. Images were acquired using an ultra-high-resolution mode (slice thickness: 0.25 mm; matrix, 1024 × 1024). Two radiologists independently evaluated the length, diameter, and intra-vestibular protrusion of the prosthesis. Post-operative air-bone gaps (ABGs) were recorded. ResultsFourteen patients were enrolled (mean age, 44.3 ± 13.8 [SD] years, 9 females), resulting in 16 temporal bone UHR-CTs. The exact length was obtained in 81.3 % (n = 13/16) and underestimated by 0.1 to 0.3 mm in the remaining 18.7 % (n = 3/16) CT scans for both readers (mean misestimation: −0.02 ± 0.06 [SD] mm, overall underestimation of 0.43 %). The exact diameter was reported in 75 % (n = 12/16) and 87.5 % (n = 14/16) of the CT scans for readers 1 and 2, respectively, and was off by 0.1 mm in all discrepancies (mean misestimation: 0.01 ± 0.04 [SD] mm, overall overestimation of 2.43 %). Intravestibular prosthesis protrusion was of 0.5 ± 0.43 [SD] mm (range: 0–1) and 0.49 ± 0.44 [SD] mm (range: 0–1.1) for readers 1 and 2, respectively, and did not correlate with ABGs (r = 0.25 and 0.22; P = 0.39 and 0.47 for readers 1 and 2, respectively). Intra and interobserver agreements were excellent. ConclusionUHR-CT provides 99.6 % and 97.6 % accuracy for prosthesis length and diameter measurements, respectively.
Read full abstract