Objective:To compare the effects of 3 implant designs, with and without hydroxyapatite reinforcement, on push/pull-out strength and rotational torque.Methods:Three implant designs (Gamma 3, INTERTAN, and PFNA-II) were selected for comparison. A hydroxyapatite cylinder (NEOBRACE) was used to reinforce the interface between the femoral head and the lag screw. Maximum push-out strength, maximum pull-out strength, and peak rotational torque were measured in cellular blocks mimicking osteoporotic cancellous bone, with and without NEOBRACE.Results:In the push-out test, INTERTAN produced a significantly higher push-out strength in osteoporotic bone density cellular blocks than the other lag screws and blades (P < .05). With the addition of NEOBRACE, push-out strength was significantly higher for INTERTAN and PFNA-II (P < .05) than for the non-NEOBRACE group. In the pull-out test, INTERTAN produced a significantly higher pull-out strength in the osteoporotic bone density cellular blocks than did the other lag screws and blades (P < .05). With the addition of NEOBRACE, the pull-out strengths of INTERTAN and Gamma 3 versus those of the non-NEOBRACE group significantly increased (P < .05). In the rotational torque test, INTERTAN produced significantly greater rotational torque in the osteoporotic cellular blocks than the other lag screws and blades (P < .05). The addition of NEOBRACE resulted in a significant increase in rotational torque only for INTERTAN (P < .05).Conclusion:The use of NEOBRACE supported an increase in push/pull-out strength and rotational torque, especially in systems with a relatively increased bone or implant interface area.Level of Evidence: Level V