A high-purity germanium (HPGe) gamma-ray spectrometer and an Alpha-GUARD radon monitor are the activity and activity concentration transfer standards of 222Rn, respectively. The gaseous 222Rn standard source traceable to the NIM's absolute standardization apparatus of 222Rn was developed for calibrating these transfer standards. The calibration results of HPGe detector at three different distances show that the uncertainties of experimental efficiency are less than 0.77% (k = 1). Meanwhile, the Monte Carlo (MC) simulation was carried out at wider distances range for the HPGe calibration as well, and the deviation at same three distances between experiments and simulations results is within ±3.0%. The advantage of calibration using MC simulation and the possible reasons for the deviation were discussed in detail in this article. In addition, a reference activity concentration environment based on the gaseous 222Rn standard source was also developed to calibrate the radon monitor, and the typical uncertainty for calibration factor of Alpha-GUARD radon monitor could be reduced to 1.9% (k = 1).