A method to monitor glucose in whole blood is presented. The aim of the project was to develop a prototype for a bedside monitor system for semi-continuous monitoring of the blood glucose concentration, requiring only one calibration. This was made possible by using the special advantage of the thermal sensor technique in combination with the adjustment of flow. The glucose concentration was determined from the difference between the sensor response and an estimated background signal. Using standard addition technique, calibration factors for background and sensitivity were set and remained unchanged during the monitoring. The background signal was 45 ± 8 mV (mean ± S.D., n = 8) and the sensitivity was 28 ± 1 mV/mmol (mean ± S.D., n = 4). Recovery in whole blood was 90–98% (mean 94%, n = 12). With an injection interval of 3 min the precision with the sensor was <3% over more than 100 blood samples. Response time was about 60 s. The calculated glucose values correlated, r = 0.98, with the values obtained with an YSI glucose; analyser (Yellow Springs Instruments, Yellow Springs, OH, USA), over the range 2–20 mmol/l.