Historically, pressure limited, time cycled ventilation has been the cornerstone of neonatal ventilation. The advent of sensitive flow sensors now allows rapid, continuous measurement of Vt in small neonates. We hypothesized that an algorithm combining Vt measurements with maximal use of noninvasive monitoring would allow safe, uniform weaning of infants with RDS from mechanical ventilation with a minimum of blood draws. The primary premise of the algorithm: using Vt to modify peak inspiratory pressure (PIP) to keep Vt 5-7 ml/kg; modify rate to keep pH 7.27-7.33; modify FiO2 to maintain SaO2 89-94%. Tidal volumes were measured with the Bicore CP-100 Neonatal monitor. This uses a self calibrated transducer to measure air flow. Flow is then integrated to give the infant's tidal volume. CO2 was monitored with transcutaneous CO2 (Novametrix), SaO2 with the Nellcor pulse oximeter. Arterial blood gases were drawn from the umbilical arterial line when available, or peripheral capillary blood gases were followed. Data from 33 infants (mean BW=1326g, GA=29.5 wks, 54% male, 50% Caucasian, initial mean airway pressure (MAP)=8.5 cmH2O, initial FIO2= 0.62) intubated secondary to RDS and managed as per the WP were evaluated. This was compared with data from 33 matched retrospective controls (mean BW=1274 g, GA=29.2 wks, 57% male, 66% Caucasian, initial MAP=8.7 cm H2O, FIO2=0.63). Primary outcome variables included: initial PIP (IPIP) and maximum PIP (MPIP), total blood gases (BG) for initial 72 hours of management and for first 7 days of hospitalization, and duration of mechanical ventilation. No patient managed per the weaning protocol had a pH value of <7.25. There was a significant difference between the two groups in terms of IPIP (WP=21cm H2O vs NO WP= 24 cm H2O, p<0.03), MPIP (WP = 21.8 cm H2O vs NO WP= 25.0 cmH2O, p<0.01). The infants managed with the WP had an average of 13.6 BGs drawn during the first 72 hours vs. 22.5 BGs for the NO WP (p<0.001). For the first week, the WP infants had 19.0 BG vs. 33.5 BG (p < 0.01). In terms of hospital charges for BG in the first week, the WP charges were $1452 vs $2560 for the NO WP. Actual hospital COST (therapist time and supplies per blood gas) resulted in saving $65 per patient for the first week ($85 vs. $150; p<0.01). Although the infants on the WP had a trend for decreased time on ventilation, there was no significant difference. Institution of the Vt driven WP enhanced our ability to wean comparable infants with RDS in a uniform manner with significantly decreased IPIP and MPIP, and decreased BGs at both 72 hours and 7 days. We hypothesize that Vt driven WP may decrease barotrauma, and with their concurrent savings, can safely provide consistent and cost effective weaning of infants with acute RDS.
Read full abstract