PurposeThe aim was to investigate the variation of the arterial input function (AIF) within and between various DCE MRI sequences. Material and methodsA dynamic flow-phantom and steady signal reference were scanned on a 3T MRI using fast low angle shot (FLASH) 2d, FLASH3d (parallel imaging factor (P) = P0, P2, P4), volumetric interpolated breath-hold examination (VIBE) (P = P0, P3, P2 × 2, P2 × 3, P3 × 2), golden-angle radial sparse parallel imaging (GRASP), and time-resolved imaging with stochastic trajectories (TWIST). Signal over time curves were normalized and quantitatively analyzed by full width half maximum (FWHM) measurements to assess variation within and between sequences. ResultsThe coefficient of variation (CV) for the steady signal reference ranged from 0.07-0.8%. The non-accelerated gradient echo FLASH2d, FLASH3d, and VIBE sequences showed low within sequence variation with 2.1%, 1.0%, and 1.6%. The maximum FWHM CV was 3.2% for parallel imaging acceleration (VIBE P2 × 3), 2.7% for GRASP and 9.1% for TWIST. The FWHM CV between sequences ranged from 8.5–14.4% for most non-accelerated/accelerated gradient echo sequences except 6.2% for FLASH3d P0 and 0.3% for FLASH3d P2; GRASP FWHM CV was 9.9% versus 28% for TWIST. ConclusionMRI acceleration techniques vary in reproducibility and quantification of the AIF. Incomplete coverage of the k-space with TWIST as a representative of view-sharing techniques showed the highest variation within sequences and might be less suited for reproducible quantification of the AIF
Read full abstract