PurposeTo investigate the feasibility of utilizing the Sphinx Compact detector for quality assurance in a uniform scanning proton therapy system. MethodThe Sphinx Compact detector was used to measure various dosimetric parameters of uniform scanning proton beam at the Oklahoma Proton Center: distal range, distal-fall-off, collinearity, field symmetry, flatness, and field size for four different beams. A specially designed brass aperture was used to perform the required measurements. The Sphinx Compact measurement results were validated against the measurement results from the well-established detectors in proton therapy: IBA Zebra, IBA MatriXX-PT, EBT3 films, and Logos XRV-124. The data collected using the Sphinx Compact was analyzed in myQA software. ResultsBased on the data analysis performed, the Sphinx Compact measurements were within acceptable accuracy to the results from the detectors mentioned in the Method section. Specifically, the lateral penumbra was within ±0.4 mm, collinearity was within ± 0.5 mm, flatness was within ±0.6 %, symmetry within ±1.6 %, distal range was within ±0.5 mm, distal-fall-off was <0.9 mm, and field size was within ±1 mm. The reproducibility of the Sphinx Compact was tested for range and collinearity, and the results were within ±0.1 mm. ConclusionThe sphinx Compact detector could potentially replace multiple detectors utilized for monthly QA in uniform scanning proton therapy. In a multi-room center, performing the QA with one detector compared to using multiple detectors dramatically reduces total QA time and the complexity of the QA process.