Due to its important environmental role, the analysis of trace metals in natural waters is attracting increasing attention; consequently, faster and more accurate analytical methods are now needed to reach even lower limits of detection. In this work, we propose the use of electro-membrane extraction (EME) to improve analytical methods based on hollow fiber liquid phase micro-extraction (HFLPME). Specifically, an EME-based method for the determination of trace Ni in seawater has been developed, using an HFLPME system with di-2-ethylhexyl phosphoric acid (DEHPA) in kerosene as a chemical carrier, followed by instrumental determination by graphite furnace atomic absorption spectroscopy (GFAAS). Under optimum conditions, Ni was pre-concentrated 180 ± 17 times after 15 min, using sample pH = 5.5, the concentration of DEHPA 0.9 M in the liquid membranes, and 1.9 M HNO3 in the acceptor solution, as well as an electric potential of 25 V with the sample being stirred at 500 rpm. When compared with other HFLPME systems for pre-concentration of trace Ni in seawater in the absence of electric potential, the enrichment factor was improved 2.2 times, while the time of extraction was reduced an 89%. The limit of detection of the new method was 23.3 ng L-1, and both its applicability and accuracy were successfully evaluated by analyzing Ni concentration in a seawater-certified reference material (BCR-403), showing the reliability of EME for sample preparation in the determination of trace metals in marine water samples.