To develop and standardize a reliable in vitro dynamic thrombogenicity test protocol, the key test parameters that could impact thrombus formation need to be investigated and understood. In this study, we evaluated the effect of temperature on the thrombogenic responses (thrombus surface coverage, thrombus weight, and platelet count reduction) of various materials using an in vitro blood flow loop test system. Whole blood from live sheep and cow donors was used to assess four materials with varying thrombogenic potentials: negative-control polytetrafluoroethylene (PTFE), positive-control latex, silicone, and high-density polyethylene (HDPE). Blood, heparinized to a donor-specific concentration, was recirculated through a polyvinyl chloride tubing loop containing the test material at room temperature (22-24°C) for 1 hour, or at 37°C for 1 or 2 hours. The flow loop system could effectively differentiate a thrombogenic material (latex) from the other materials for both test temperatures and blood species ( p < 0.05). However, compared with 37°C, testing at room temperature appeared to have slightly better sensitivity in differentiating silicone (intermediate thrombogenic potential) from the relatively thromboresistant materials (PTFE and HDPE, p < 0.05). These data suggest that testing at room temperature may be a viable option for dynamic thrombogenicity assessment of biomaterials and medical devices.