The Thompson hemiarthroplasty is a common treatment option for acute neck of femur fractures in the elderly population. Our department noted a significant number of patients returning with thigh pain, radiographic loosening and femoral osteolysis following cemented implantation of the titanium alloy version of the Thompson hemiarthroplasty. We are not aware of any previous reports documenting complications specific to the titanium Thompson implant and a retrospective cohort study was therefore initiated following clinical governance approval.366 titanium alloy Thompson prostheses were implanted for hip fracture treatment between 2017 and 2020. As of February 2023, 6 of these have been revised at our hospital. 5 were revised for symptomatic femoral osteolysis and 1 presented with an acute periprosthetic fracture. All revised cases were determined to be aseptic. 32 living patients were excluded from recall on compassionate grounds due to permanent nursing home residence. 47 living patients were identified of which 33 attended for xray. 28 deceased and/or nursing home resident patients who had pelvis x-rays in the previous 12 months were also included in the analysis. Including the 6 index hips already revised, a total of 61 hip xrays were analysed, of which 19 hips (31.1%) showed radiographic evidence of femoral osteolysis or loosening.We conclude that there is a concerning incidence of femoral osteolysis and implant loosening associated with the titanium Thompson implant. We have discontinued use of the implant and reported our experience to the MHRA. We encourage other Scottish Health-Boards who use this implant to consider enhanced follow-up.