Background and purposeIn neuro-oncology, high spatial accuracy is needed for clinically acceptable high-precision radiation treatment planning (RTP). In this study, the clinical applicability of anatomically optimised 7-Tesla (7T) MR images for reliable RTP is assessed with respect to standard clinical imaging modalities.Materials and methodsSystem- and phantom-related geometrical distortion (GD) were quantified on clinically-relevant MR sequences at 7T and 3T, and on CT images using a dedicated anthropomorphic head phantom incorporating a 3D grid-structure, creating 436 points-of-interest. Global GD was assessed by mean absolute deviation (MADGlobal). Local GD relative to the magnetic isocentre was assessed by MADLocal. Using 3D displacement vectors of individual points-of-interest, GD maps were created. For clinically acceptable radiotherapy, 7T images need to meet the criteria for accurate dose delivery (GD < 1 mm) and present comparable GD as tolerated in clinically standard 3T MR/CT-based RTP.ResultsMADGlobal in 7T and 3T images ranged from 0.3 to 2.2 mm and 0.2–0.8 mm, respectively. MADLocal increased with increasing distance from the isocentre, showed an anisotropic distribution, and was significantly larger in 7T MR sequences (MADLocal = 0.2–1.2 mm) than in 3T (MADLocal = 0.1–0.7 mm) (p < 0.05). Significant differences in GD were detected between 7T images (p < 0.001). However, maximum MADLocal remained ≤1 mm within 68.7 mm diameter spherical volume. No significant differences in GD were found between 7T and 3T protocols near the isocentre.ConclusionsSystem- and phantom-related GD remained ≤1 mm in central brain regions, suggesting that 7T MR images could be implemented in radiotherapy with clinically acceptable spatial accuracy and equally tolerated GD as in 3T MR/CT-based RTP. For peripheral regions, GD should be incorporated in safety margins for treatment uncertainties. Moreover, the effects of sequence-related factors on GD needs further investigation to obtain RTP-specific MR protocols.