Our purpose in this study was to evaluate the performance of four-dimensional computed tomography (4D-CBCT) and to optimize the acquisition parameters. We evaluated the relationship between the acquisition parameters of 4D-CBCT and the accuracy of the target motion trajectory using a dynamic thorax phantom. The target motion was created three dimensionally using target sizes of 2 and 3cm, respiratory cycles of 4 and 8s, and amplitudes of 1 and 2cm. The 4D-CBCT data were acquired under two detector configurations: "small mode" and "medium mode". The projection data acquired with scan times ranging from 1 to 4min were sorted into 2, 5, 10, and 15 phase bins. The accuracy of the measured target motion trajectories was evaluated by means of the root mean square error (RMSE) from the setup values. For the respiratory cycle of 4s, the measured trajectories were within 2mm of the setup values for all acquisition times and target sizes. Similarly, the errors for the respiratory cycle of 8s were <4mm. When we used 10 or more phase bins, the measured trajectory errors were within 2mm of the setup values. The trajectory errors for the two detector configurations showed similar trends. The acquisition times for achieving an RMSE of 1mm for target sizes of 2 and 3cm were 2 and 1min, respectively, for respiratory cycles of 4s. The results obtained in this study enable optimization of the acquisition parameters for target size, respiratory cycle, and desired measurement accuracy.