IntroductionThe aims of the study: (1) to compare the Super Pulse Thulium Fiber Laser (SP TFL) and the holmium: yttrium–aluminium-garnet (Ho:YAG) lasers in retrograde intrarenal surgery (RIRS); (2) to compare the efficacy of SP TFL laser fibers of different diameters (150 μm and 200 μm).MethodsA prospective randomized single-blinded trial was conducted. Patients with stones from 10 to 20 mm were randomly assigned RIRS in three groups: (1) SP TFL (NTO IRE-Polus, Russia) with fiber diameter of 150 μm; (2) SP TFL with 200-μm fiber; and (3) Ho:YAG (Lumenis, USA) with 200-μm fiber.ResultsNinety-six patients with kidney stones were randomized to undergo RIRS with SP TFL using a 150-μm fiber (34 patients) and a 200-μm fiber (32 patients) and RIRS with Ho:YAG (30 patients). The median laser on time (LOT) in the 200-μm SP TFL group was 9.2 (6.2–14.6) min, in 150-μm SP TFL—11.4 (7.7–14.9) min (p = 0.390), in Ho:YAG—14.1 (10.8–18.1) min (p = 0.021). The total energy consumed in 200-μm SP TFL was 8.4 (5.8–15.2) kJ; 150-μm SP TFL − 10.8 (7.3–13.5) kJ (p = 0.626) and in Ho:YAG—15.2 (11.1–25.3) kJ (p = 0.005).ConclusionsIrrespective of the density, RIRS with SP TFL laser has proven to be both a safe and effective procedure. Whilst the introduction of smaller fibers may have the potential to reduce the duration of surgery, SP TFL results in a reduction in the LOT and total energy for stone ablation in RIRS compared with Ho:YAG.