The goal of this study was to test the static and dynamic strength and loosening resistance of the posterior flange of a novel total elbow arthroplasty. We also examined the forces experienced by the ulnohumeral joint and the posterior olecranon during expected elbow use. Static stress analysis was performed for 3 flange sizes. Failure testing was conducted on 5 flanges (1 medium size and 4 small sizes). Loading occurred to reach 10,000 cycles. If this was accomplished, the cyclic load was increased until failure occurred. If failure occurred before 10,000 cycles, a lower force was employed. The safety factor for each implant size was calculated, and implant failure or loosening was observed. Static testing revealed a safety factor of 6.6, 5.74, and 4.53 for the small, medium, and large flanges, respectively. The medium-sized flange completed 10,000 cycles with 1,000 N at 1 Hz, and then the force was increased until it failed at 23,000 cycles. Two small-sized flanges failed at 2,345 and 2,453 cycles, respectively, when loaded with 1,000 N. Two more small flanges were loaded with 729 N for 10,000 cycles, and then the cyclic load was continued until they failed at 17,000 and 17,340 cycles, respectively. No screw loosening was noted in any specimens. This study demonstrates that the posterior flange withstood static and dynamic forces greater than what is expected during invivo use of a novel total elbow arthroplasty design. Static strength calculation and cyclic loading demonstrate that the medium-sized posterior flange is stronger than the small-sized posterior flange. Ensuring that the ulnar body component and the posterior flange maintain secure connectivity with the polyethylene wear component may be beneficial to the proper function of a novel nonmechanically linked total elbow arthroplasty.