Gerresheimer and Midas Pharma have developed a novel cartridge-based autoinjector concept in which the cartridge as primary packaging is under constant pressure. In this article standard cartridge primary packaging material of five different companies were analyzed for their behavior under long-term pressure. Materials of 3 glass manufacturers and 2 manufacturers for cartridge rubber parts were considered. Within the test program septum stability, septum piercing, glide forces (GF), break-loose forces (BLF), glass breaking as well as a regulatory approved and marketed antibody drug product under pressure were subject to analysis. Under pressure the cartridge septum bulge grew within the first 14 days and then relevantly slowed down. An accelerated study in different atmospheric conditions allowed to extrapolate values for 24 months storage, not showing any signs of decay or problematic septum bulge increase. Pierce forces were in normal ranges and septum rupture could not be observed at the end of 42 days of pressurization. GF and BLF were within acceptable ranges and changes due to pressure could not be observed. Lowest glass breaking pressures at 4922kPa turned out to be at least 3.5 times higher than pressures used in the autoinjector concept. Degradation of the Adalimumab antibody drug product due to pressure or device fluid pathway could not be observed with size exclusion chromatography, electrophoresis or sub-visible particles tested as a release testing in a GMP setting.