To confirm the performance improvement of virtual monoenergetic images (VMIs) for iodine contrast tasks in a clinical photon-counting detector CT (PCD CT) using Fourier-based assessment, compared with those in the latest-generation dual-source dual-energy CT (DECT). A water-filled bath with a diameter of 300mm, which contains rod-shaped phantoms equivalent to diluted iodine (2 and 12mg/mL), was scanned using PCD CT and DECT at 15, 7.5, and 3 mGy. VMIs were generated without any iterative reconstruction algorithm. Task transfer function (TTF), noise power spectrum (NPS), and slice sensitivity profile were evaluated for VMIs at 70 and 40keV. The detectability index (d') and the squared system performance function (SPF2) calculated by TTF2/NPS were compared. At 40keV, the d' values of PCD CT were higher (percentage increase of 25.7-39.9%) than those of DECT, whereas at 70keV, the difference was rather small. The SPF2 values at 40keV of PCD CT grew notably higher than those of DECT as the spatial frequency increased. The higher SPF2 values endorsed the lower image noise and the sharper edge of the rod phantom as observed. The d' and SPF2 in VMIs at 40keV of PCD CT were notably higher than those of DECT, which endorsed the clinical advantages of PCD CT that had been previously reported in various studies.