Nd and Sm isotope ratios play an important role in geological dating and as nuclear forensic signatures; however, the overlap of the respective 144, 148, 150 Nd/Sm isobars requires prior separations to be performed before analysis on typical MS platforms. The work presented here overcomes these isobaric interferences using ultrahigh-mass resolution to alleviate interference without prior chemical separations. A liquid sampling-atmospheric pressure glow discharge ion source was coupled to a standard, QExactive Focus Orbitrap mass spectrometer, providing a mass resolution of ~80 k. A Spectroswiss FTMS booster X2 data acquisition package was used to collect extended transients, providing much higher mass resolution; ~230 k and ~600 k are employed here for Nd and Sm isotopes. While the standard Orbitrap resolution is far greater than typical "atomic" MS platforms, it was insufficient to alleviate all isobars. The use of a resolution of ~230 k resulted in baseline separation across the entire isotopic envelope for both Nd and Sm. Isotope ratios obtained from Nd:Sm mixtures using high-resolution were equivalent to those found for individual-element solutions, while isotope ratios obtained at a resolution of ~80 k (standard for the OEM data system) showed large deviations. Use of ultrahigh-resolution is an attractive alternative to extensive chemical separations to alleviate severe isobaric interferences. Sufficient mass resolution greatly reduces/eliminates the need for sample manipulations (separations) before analysis while reducing costs and total analysis times.