The performance of columns packed with 1.7 µm particles for aqueous size-exclusion chromatography was assessed at high-pressure conditions and linked to particle- and column-bed integrity. Decreasing the particle size from 3.5 µm to 1.7 µm increases the resolution due to the improved mass-transfer characteristics, allowing to significantly speed-up analysis without compromising the selectivity. A sub-minute separation of intact proteins was realized on a 4.6 mm i.d × 75 mm long column packed with 1.7 µm SEC particles applying a flow rate of 1.8 mL/min, corresponding to a column pressure of 530 bar. Ultra-high pressure operation (exceeding manufacturer's recommendations) resulted in peak deformation, a shift towards earlier retention times, and an alteration in selectivity. To gain insights in the mechanisms of column deterioration, short 30 mm long columns were operated at UHPLC conditions, maximizing the pressure drop over individual particles. This resulted in the presence of fractured particles situated at the column outlet, as verified by scanning electron micrographs. Mercury-intrusion porosimetry and argon-adsorption measurements did not reveal significant differences in intraparticle volume between particle batches sampled before and after pressure stress testing. As particles at the column outlet fracture (but not collapse) at high pressure operation, a void was formed at the column inlet. The degradation of the separation performance appeared to be the result of a decrease in interparticle pore volume.