Human body temperature is a fundamental physiological sign that reflects the state of physical health. It is important to achieve high-accuracy detection for non-contact human body temperature measurement. In this article, a Ka band (32 to 36 GHz) analog complex correlator using the integrated six-port chip is proposed, and a millimeter-wave thermometer system based on the designed correlator is completed for human body temperature measurement. The designed correlator utilizes the six-port technique to achieve large bandwidth and high sensitivity, and miniaturization of the correlator is achieved through an integrated six-port chip. By performing the single-frequency test and the broadband noise measurement on the correlator, we can determine that the dynamic range of input power of the correlator is -70 dBm to -35 dBm, and the correlation efficiency and equivalent bandwidth are 92.5% and 3.42 GHz, respectively. Moreover, the output of the correlator varies linearly with the input noise power, which reveals that the designed correlator is suitable for the field of human body temperature measurement. Then, a handheld thermometer system, with a size of 140 mm × 47 mm × 20 mm, is proposed using the designed correlator, and the measurement results show that the temperature sensitivity of the thermometer is less than 0.2 K.