To date, only a few studies have compared radial head prostheses (RHP) and their different anchoring principles. The aim of this study was to characterize concomitant injuries, necessary adjuvant procedures, complications, and radiological findings after implantation of two different types of RHP. Sixty-six patients with radial head fractures were treated with MoPyC (Tornier/France, 50mm stem, monopolar) or SBI rHead prostheses (Small Bone Innovations/USA, 22-mm stem, bipolar) and followed up over 42months (16-64months). Primary objective was the detection of different loosening and explantation rates. In addition to the revision rate and the reasons for revision, we also used radiological findings to assess the dynamics of lysis phenomena. Thirty-five patients (mean age 48years; 22-73years) were treated with the MoPyC, and 31 patients (mean age 47years, 19-69years) with the rHead prosthesis. Of these, 98% had a Mason 3 or 4 type of radial head fracture, and 94% showed concomitant injuries, which were addressed in 89% of cases by adjuvant procedures. The surgical revision rate was 20% (rHead 23%; MoPyC 18%). Significantly more of the rHead prostheses had to be explanted compared to MoPyC prostheses because of painful loosening (16% vs. 3%; p= 0.029). Predictors of subsequent loosening were significant radiolucent lines in the RHP over the first six months and an increase in width by two and a half times within the first 50days after implantation. Complex radial head fractures are frequent and difficult to treat. The monopolar long-stemmed prosthesis showed significantly lower rates of painful loosening and explanation rates than the bipolar short-stemmed prosthesis.