A selective and sensitive method is proposed for the determination of mercury by anodic stripping voltammetry after its preconcentration from the gas phase. Mercury from the sample solution is reduced to elemental Hg by SnCl2 and volatilized by the bubbles of a carrier gas. The gas containing mercury vapour is dried and passed through a capillary onto a gold coated graphite electrode. An anodic stripping voltam-mogram is recorded from 0.1 mol/1 HClO4 + 3 × 10−3 mol/1 HCl solution. The calibration curve is linear from 1 × 10−9 to 4 × 10−8 mol/1 Hg(NO3)2. The absolute detection limit is 0.46 ng Hg. The relative standard deviations for 4 × 10−9 mol/1 and 2 × 10−8 mol/1 Hg(NO3)2 are 9.8% and 6.1%, respectively n = 5).