With the International Commission on Radiological Protection (ICRP) lowering the annual dose limit for the eye lens to 20 mSv, precise monitoring of eye lens exposure has become essential. The personal dose equivalent at a depth of 3 mm, Hp(3), is the measurement method for monitoring the dose to the lens of the eye. Traditional dosimetry methods primarily address lateral radiation exposure scenarios, where radiation approaches from the left or right, necessitating the rotation of the phantom during type testing around the vertical axis. However, these methods do not adequately account for bottom-to-top radiation exposures which are common in real-world situations (such as radiation scattered by the patient reaching medical staff).
This study examines oblique radiation exposure conditions using a typical eye lens thermoluminescent dosemeter (TLD), Eye-D, placed on a cylindrical phantom to assess dose responses at different angles and exposure energies. The study employs both lower-energy (N-30 radiation quality) and higher-energy (N-100 radiation quality) X-rays at irradiation angles of -60°, 0°, and +60°, measured along the vertical and horizontal rotation axes of the phantom.
The results show no significant differences between horizontal and vertical (polar and radial) rotation orientations of the dosemeter-phantom setup: recorded relative doses stayed well within ±1%, i.e., by far within the attributed combined uncertainty of ±2%.
.
Read full abstract