For technical and radioprotection reasons, it has become essential to develop new dosimetric tools adapted to the specificities of computed tomography (CT) to ensure precise and efficient dosimetry since the current standards are not suitable for clinical use and for new CT technological evolution. Thanks to its many advantages, plastic scintillating fibers (PSF) is a good candidate for more accurate and personalized real-time dosimetry in computed tomography, and the company Fibermetrix has developed a new device named IVISCAN® based on this technology. In this study, we evaluated performances of IVISCAN® and associated uncertainties in terms of dose-rate dependence, angular dependence, stability with cumulative dose, repeatability, energy dependence, length dependence, and special uniformity in reference and clinical computed tomography beam qualities. For repeatability, the standard deviation is less than 0.039%, and the absolute uncertainty of repeatability lies between 0.017% and 0.025%. The deviation between IVISCAN® and the reference regarding energy dependence is less than 1.88% in clinical use. Dose rate dependence results show a maximum deviation under ±2%. Angular dependence standard deviation σ is 0.8%, and the absolute uncertainty was 1.6%. We observed 1% of variation every 50 Gy steps up to a cumulative dose of 500 Gy. Probe response was found to be independent of the PSF length with a maximum deviation < 2.7% between the IVISCAN® probe and the 1 cm PSF probe. The presented results demonstrated that IVISCAN® performances are in accordance with metrology references and the international standard IEC61674 relative to dosemeters used in X-ray diagnostic imaging and then make it an ideal candidate for real-time dosimetry in CT applications.