To reduce radiation dose and subsequent risks, several legislative documents in different countries describe the need for Diagnostic Reference Levels (DRLs). Spinal radiography is a common and high-dose examination. Therefore, the aim of this work was to establish the DRL for Computed Tomography (CT) examinations of the spine in healthcare institutions across Jordan. Data was retrieved from the picture archiving and communications system (PACS), which included the CT Dose Index (CTDI(vol)) and Dose Length Product (DLP). The median radiation dose values of the dosimetric indices were calculated for each site. DRL values were defined as the 75th percentile distribution of the median CTDI(vol) and DLP values. Data was collected from 659 CT examinations (316 cervical spine and 343 lumbar-sacral spine). Of the participants, 68% were males, and the patients' mean weight was 69.7 kg (minimum = 60; maximum = 80, SD = 8.9). The 75th percentile for the DLP of cervical and LS-spine CT scans in Jordan were 565.2 and 967.7 mGy.cm, respectively. This research demonstrates a wide range of variability in CTDI(vol) and DLP values for spinal CT examinations; these variations were associated with the acquisition protocol and highlight the need to optimize radiation dose in spinal CT examinations.