LNE-LNHB is involved in a European project aiming at establishing absorbed dose-to-water standards for photon-radiation fields down to 2 × 2 cm2. This requires the calibration of reference ionization chambers of small volume. Twenty-four ionization chambers of eight different types with volume ranging from 0.007 to 0.057 cm3 were tested in a 60Co beam. For each chamber, two major characteristics were investigated: (1) the stability of the measured current as a function of the irradiation time under continuous irradiation. At LNE-LNHB, the variation of the current should be less than ±0.1% in comparison with its first value (over a 16 h irradiation time); (2) the variation of the ionization current with the applied polarizing voltage and polarity. Leakage currents were also measured. Results show that (1) every tested PTW (31015, 31016 and 31014) and Exradin A1SL chambers demonstrate a satisfying stability under irradiation. Other types of chambers have a stability complying with the stability criterion for some or none of them. (2) IBA CC01, IBA CC04 and Exradin A1SL show a proper response as a function of applied voltage for both polarities. PTW, Exradin A14SL and Exradin A16 do not. Only three types of chambers were deemed suitable as reference chambers according to LNE-LNHB requirements and specifications from McEwen (2010 Med. Phys. 37 2179–93): Exradin A1SL chambers (3/3), IBA CC04 (2/3) and IBA CC01 (1/3). The Exradin A1SL type with an applied polarizing voltage of 150 V was chosen as an LNE-LNHB reference chamber type in 2 × 2 cm2 radiation fields.