Left ventricular volume has been measured with ultrafast computed tomography. However, the accuracy with which this can be done is unknown. We therefore imaged with ultrafast computed tomography 11 rectangular phantoms, 20 to 225 ml, and 17 left ventricular casts, 15 to 112 ml. Two observers planimetered serial tomographic images and computed volume from sequential tomograms. There was no significant inter- or intraobserver difference in measurement of phantoms. Deviation of ultrafast computed tomographic volume from true phantom volume was -0.1 +/- 3.5% SD, range 9.0 to -7.6%. Correlation of true phantom volume with ultrafast computed tomographic volume was 0.99, SEE = 1.9 ml. No significant difference was observed between merged and single ultrafast computed tomographic scanning sequences. Left ventricular cast volume determined by ultrafast computed tomography deviated from true volume by 6% +/- 20%, range 54% to -45%. Correlation of true volume with ultrafast computed tomographic volume was 0.99, SEE = 5.1 ml. There was no interobserver significant difference in measurement of left ventricular cast volume. Correlation between ultrafast computed tomographic volume and cineradiographic volume of the same left ventricular casts was 0.99, SEE = 4.4 ml. Thus, phantom volumes can be measured accurately without significant intra- or interobserver variation. Merged scanning sequences did not influence volume determination. Left ventricular cast volume determination was comparable to that obtained with cineradiography.