PurposeTo investigate the effect of simultaneous use of dual applicators on the image quality of real-time magnetic resonance (MR) thermometry and to characterize the dual-applicator treatment zone pattern during MR imaging–guided microwave ablation (MWA). Materials and MethodsMWA experiments were performed on a 1.5-T MR scanner with 2 commercial microwave systems (902–928 MHz). Phantom experiments were first performed to evaluate the effect of dual-applicator MWA on the image quality of MR. Then, porcine tissue model experiments were conducted with real-time MR thermometry using either a single applicator or dual applicators inserted 2.6, 3.6, and 4.6 cm apart. Fiberoptic thermal probes were used to measure the temperature changes at the tissue surface. ResultsSimultaneous use of dual applicators resulted in a decrease in the relative signal-to-noise ratio (SNR) in the MR thermometry images to 55% ± 2.9% when compared with that of a single applicator (86.2% ± 2.0%). Despite the lower SNR, the temperature and ablation zone maps were of adequate quality to allow visualization of the ablation zone(s). The extents of increase in the temperature at the tissue surface using dual applicators (19.7 °C ± 2.6 °C) and a single applicator (18.2 °C ± 3.3 °C) were not significantly different (P = .40). Treatment zones were significantly larger (P < .05) in dual-applicator ablations (29.4 ± 0.4, 39.9 ± 0.6, and 42.6 ± 0.9 cm2 with 2.6-, 3.6-, and 4.6-cm spacing, respectively) at the end of the ablation procedure than in the single-applicator MWA (18.6 ± 0.9 cm2). ConclusionsMR imaging–guided dual-applicator MWA produced larger ablation zones while allowing adequate real-time MR thermometry image quality for monitoring the evolution of the treatment zone.