Intraoral coils (IOCs) in magnetic resonance imaging (MRI) significantly improve the signal-to-noise ratio compared with conventional extraoral coils. To assess the safety of IOCs, we propose a 2-step procedure to evaluate radiofrequency-induced heating of IOCs and compare maximum temperature increases in 3 different types of IOCs. The 2-step safety assessment consists of electric field measurements and simulations to identify local hotspots followed by temperature measurements during MRI. With this method, 3 different coil types (inductively coupled IFC, transmit/receive tLoop, and receive-only tLoopRx) were tested at 1.5 T and 3 T for both tuned and detuned coil states. High SAR and regular MRI protocols were applied for 2 coil positions. The measured E field maps display distinct hotspots for all tuned IOCs, which were reduced by at least 40-fold when the IOCs were detuned. Maximum temperature rise was higher when the coils were positioned at the periphery of the phantom with the coil planes parallel to B0. When neither active nor passive detuning was applied, maximum temperature increase of ΔT = 1.3/0.5/1.8 K was found for IFC/tLoop/tLoopRx coils. Hotspots detected by E field measurements, and simulations were consistent. In the simulations, the results were different for homogeneous phantoms compared with full anatomical models. The 2-step test procedure is applicable to different coil types. The results indicate that a risk for radiofrequency-induced heating exists for tuned IOCs, so that adequate detuning circuits need to be integrated in the coils to ensure safe operation.