Suture anchor failures can lead to revision surgeries which are costly and burdensome for patients. The durability of musculoskeletal reconstructions is therefore partly affected by the design of the suture anchors. The purpose of the study was to quantify the strength of different suture anchors whose sizes are suitable for attaching artificial Achilles and tibialis cranialis tendons in a rabbit model, as well as determine the effect of cyclic loading on the anchoring strength. Four anchors (two with embedded eyelet and two with raised eyelet, n=5 per group) were tested with cyclical loading (1000 cycles and 4.5 mm/sec) and without cycling, to inform the failure loads and mode of failure of the suture anchors. An eyebolt screw with smooth eyelet was used as a control for the test groups. All samples in all groups completed 1000 cycles and failed via suture breakage in both test conditions. All anchors had failure loads exceeding the peak Achilles tendon force in rabbits during hopping gait. The data analysis showed an effect of anchor type on the maximum tensile force at failure (F max ) in all suture categories but not an effect of loading condition. Also, the Anika anchor had a significantly less adverse effect on suture strength compared to Arthrex anchor (p=0.015), IMEX anchor (p=0.004) and Jorvet anchor (p<0.001). We observed a greater percentage of failure at the mid-section for the anchors with the raised eyelets compared to the anchors with embedded eyelets, which all failed at the knot. Anchors with embedded eyelets had clinically preferred mode of failure with less adverse effects on suture and, may be more reliable than anchors with raised eyelets for attaching artificial Achilles and tibialis cranialis tendons in rabbits.