Abstract We report the commissioning results of Elekta Unity for the dosimetric performance and mechanical quality assurance (QA), and propose additional commissioning procedures. Mechanical tests included multi-leaf collimator (MLC) positional accuracy, radiation isocenter diameter at the center and off-center position, and coincidence between the magnetic resonance (MR) image center and radiation isocenter. Comparisons between the measurements and calculations of the simple irradiated field, intensity modulated radiation therapy (IMRT) commissioning, MLC output factor ratio, validation of independent dose calculation software and end-to-end testing were performed to evaluate dosimetric performance. The average values of the MLC positional accuracy for film- and imaging device-based analysis were −0.1 and 0.3 mm, respectively. The measured radiation isocenter size was 0.41 mm, and the off-center results were within 1 mm. The coincidence was −0.21, −1.19 and 0.49 mm along the x-, y- and z-axes, respectively. The calculated percent depth doses (PDD) and profiles agreed with the measurements. The results of independent dose calculation were within the action level recommended by American Associations of Physicist in Medicine. The gamma passing rate (GPR) for IMRT commissioning was 98.6 ± 0.9%, and end-to-end testing of adapted plans showed agreement within 2% between the measurement and calculation. We reported the results of mechanical and dosimetric performances of Elekta Unity, and proposed novel commissioning procedures. Our results should provide knowledge to the physics community for enhancing the QA programs.