A likely consequence of the discontinued distribution and sale of the HVAD System (Medtronic, Minneapolis, MN) will be an increase in replacement with the HeartMate 3 (Abbott, Chicago, IL) left ventricular assist device when device exchange is necessary. If part or all of the HVAD 10-mm-diameter outflow graft is retained during replacement, the HeartMate 3 will have to run at a higher speed than it would with its 14-mm-diameter graft. A steady-state, invitro study was run with 250-mm-long samples of HVAD, HeartMate 3, and half-HVAD/half-HeartMate 3 grafts and additionally 125- and 375-mm-long samples of the HVAD graft. Flows of 3.0, 3.9, 4.3, 4.7, and 6.0 L/min were applied to encompass expected clinical conditions. At typical and high flow rates of 4.3 and 6.0 L/min, HeartMate 3 rotor speeds with the full HVAD graft had to be increased relative to those with the HeartMate 3 graft from 5350 to 5700 and 6350 to 6900 rpm, respectively, with power consumption increases from 3.7 to 4.3 W (16%) and 5.5 to 6.8 W (24%), respectively. The study did not elucidate a severe consequence of using a remnant HVAD graft during pump exchange, but the incremental risks of a higher rotor speed, disadvantage to the patient in battery runtime, and the general benefit of complete conversion to the HeartMate 3 graft should be balanced against other procedural considerations. Complete graft replacement during HVAD-to-HeartMate 3 conversion remains the preferred approach from an engineering point of view.