In patients supported by the HeartMate 3 left ventricular assist device (HM3 LVAD), pump speed adjustments may improve hemodynamics. We investigated the hemodynamic implications of speed adjustments in HM3 recipients undergoing hemodynamic ramp tests. Clinically stable HM3 recipients who underwent routine invasive hemodynamic ramp tests between 2015 and 2022 at our center were included. Filling pressure optimization, defined as central venous pressure (CVP) <12 mm Hg and pulmonary capillary wedge pressure (PCWP) <18 mm Hg, was assessed at baseline and final pump speeds. Patients with optimized pressures were compared to nonoptimized patients. Overall 60 HM3 recipients with a median age of 62 years (56, 71) and time from LVAD implantation of 187 days (124, 476) were included. Optimized filling pressures were found in 35 patients (58%) at baseline speed. Speed was adjusted in 84% of the nonoptimized patients. Consequently, 39 patients (65%) had optimized pressures at final speed. There were no significant differences in hemodynamic findings between baseline and final speeds ( p > 0.05 for all). Six and 12 month readmission-free rates were higher in optimized compared with nonoptimized patients ( p = 0.03 for both), predominantly due to lower cardiac readmission-free rates ( p = 0.052). In stable outpatients supported with HM3 who underwent routine ramp tests, optimized hemodynamics were achieved in only 2 of 3 of the patients. Patients with optimized pressures had lower all-cause readmission rates, primarily driven by fewer cardiac-related hospitalizations.