To determine and report the underlying cause of local inflammation causing recurrent neuropathy and multiple operations in a patient with a Barricaid® device. After removal of this patient's Barricaid® device, we sent local inflammatory tissue to pathology for histochemical analysis. Upon discovery of giant cells formation with polarizable foreign bodies, we performed a literature review regarding the Barricaid® device and its elements. After two previous operations and three trials of conservative management, the presented patient underwent an L5/S1 TLIF with removal of her previously installed Barricaid® device. There were no signs of device instability/failure nor were there obvious signs of infection. Inflamed tissue proximal to the Barricaid® device was discovered, debrided, and sample sent to pathology. Removal of the Barricaid® device led to subsequent and durable relief of her symptoms. During review of this case, we discovered the polyethylene terephthalate (PET) weave used in the Barricaid® device is known to induce foreign body reactions, and this precise finding was seen in the majority of animal data submitted to the FDA for the device's acceptance. Given the constellation of this patient's symptoms, imaging, intraoperative, and pathology findings, previously published reports, and pre-approval data submitted to the FDA, we conclude that the inflammatory response to the PET weave in this patient's Barricaid® device was the ultimate cause of her continued neuropathy despite multiple prior surgical interventions.