Purpose:To determine the agreement between intraocular pressure (IOP) measurements using an automated non-contact tonometer (NCT), Goldmann applanation tonometer (GAT), and the ocular response analyzer (ORA) in subjects with primary congenital glaucoma (PCG).Methods:Twenty-nine eyes of 17 PCG patients underwent IOP measurements using NCT, GAT and ORA. Variables obtained by the ORA were corneal-compensated IOP (IOPcc), Goldmann-correlated IOP (IOPg), corneal hysteresis (CH), and corneal resistance factor (CRF). A difference more than 1.5 mmHg for IOP was considered as clinically relevant.Results:Mean age of the patients was 12 years. Mean IOP (±standard deviation, SD) was 15.3 ± 2.8 mmHg (GAT), 15.5 ± 6.0 (NCT), 19.2 ± 7.0 (IOPg), and 21.1 ± 7.9 (IOPcc); (P = 0.001). Except for NCT vs. GAT (P = 1.0), the average IOP difference between each pair of measurements was clinically relevant. The 95% limits of agreements were − 10.2 to 10.3 mmHg (NCT vs. GAT), −7.8 to 15.3 (IOPg vs. GAT), and − 8.1 to 19.0 (IOPcc vs. GAT). The differences in IOP measurements increased significantly with higher average IOP values (r = 0.715, P = 0.001, for NCT vs. GAT; r = 0.802, P < 0.001, for IOPg vs. GAT; and r = 0.806, P < 0.001, for IOPcc vs. GAT). CH showed a significant association with differences in IOP measurements only for IOPcc vs. GAT (r = 0.830, P < 0.001).Conclusion:Mean IOP obtained by NCT was not significantly different from that of GAT, but ORA measured IOPs were significantly higher than both other devices.