Basic characteristics of different batches of Radix RN15® dosimeter, a commercially available undyed polymethylmethacrylate dosimeter, was studied for its application to process control of radiation sterilization. Radix has relatively small deviation of thickness, optical absorption spectrum, and optical density values before irradiation. The response curves, in terms of net optical density per unit thickness (ΔOD/mm) as a function of absorbed dose, D PMMA, are nearly linear up to 15 kGy and become sublinear at higher doses. Scattering of dose response in 5 dosimeter replicates is ± 1% (1 σ). The dose-response slope at 25 kGy while held at irradiation temperatures in the range 0–60°C, relative to those under 25°C, increases with temperature up to 40°C, the maximum point, and decreases at higher temperatures. Over the temperature range of 10–50°C, the variation with temperature of evaluated dose derived from the calibration curve at 25°C is less than 5% when using a dose rate of 6 kGy/h. At lower dose rates (e.g. 0.7 kGy/h) the irradiation temperature dependence is negligible over this temperature range. The temperature dependence at 40 kGy is less severe than that at 25 kGy. The post-irradiation stability of dose response is less than 5% for more than 100 h after 25 kGy irradiation and storage at 0–35°C. The change in optical density of unirradiated dosimeters 3 years after manufacturing is negligible when the dosimeters are stored at controlled laboratory conditions of temperature (25°C) and relative humidity (40%). The 3-year-old dosimeters by irradiation to a dose of 25 kGy show about 2% lower response than that at 25 kGy at the beginning of the 3-year period.