Patient triggered ventilation (PTV), the delivery of positive pressure inflation initiated only by the infant's respiratory efforts, has been assessed in 32 infants of gestational age range of 24 to 36 weeks. A conventional newborn ventilator has been modified, such that in CPAP mode the manual breath control can be triggered by the infant's respiratory efforts to deliver a single positive pressure inflation. The triggering sensor is a pneumotachograph placed between the endotracheal tube and the ventilator circuit, this detects changes in flow and triggering occurs if the inspiratory flow exceeds 0.4 1/min. Inflation time during PTV mode was limited to 0.4 sec or less. Initially PTV was associated in the majority of infants with improvements in oxygenation (p<0.01) when contrasted with conventional ventilation. PTV was maintained till final extubation in 20 infants without complications. One infant only developed pneumothorax. In the remaining 11 infants, who tended to be less mature (p<0.01), PTV had to be discontinued after only a few hours. Predictors of failure of PTV at 1 hour were both a lack of improvement in oxygenation and a relatively slow triggering rate when related to gestational age. We conclude that PTV is a useful alternative for preterm neonates, but at present its use should be restricted to those more mature than 28 weeks gestational age.