11ABSTRACTObjectives: The objective of this work is to establish a bioanalytical technique for the quantification of atrasentan. The employed methodology for the detection and quantification of atrasentan in human plasma has a high degree of specificity, sensitivity, and simplicity. Methods: Shimadzu pumps were utilized in conjunction with an Autosampler Agilent Zorbax XDB C18 2.1 × 50 mm, 5 μm column employed as a stationary phase to achieve the chromatographic separation. Acetonitrile (0.1% formic acid) and 5 mM ammonium formate made up the mobile phase, which was optimized to have a 70:30% v/v ratio. Isocratic elution was employed in the experiment, with a flow rate of 0.150 mL/min. The entire cycle took 3.0 minutes, using a 10 μL injection volume. Results: The detection procedure is executed utilizing positive ion mode electrospray ionization (ESI) mass spectrometry under atmospheric pressure. For atrasentan, the precursor to product ion transitions used for quantification was m/z 511.623 to 354.15 and for the internal standard, verapamil, m/z 455.22 to 165. 02. The retention times for atrasentan and verapamil were found to be 1.68 and 0.96 minutes, respectively.Conclusion: In accordance with the recommendations of the United States Food and Drug Administration (USFDA) and EMA, it was ascertained that the remaining validation parameters remained within the specified range.