The HexLoc locking system was designed to prevent back-side wear of the polyethylene liner in the modular cementless metal-backed acetabular cup, but failed. Back-side wear was analyzed using clinico-radiological data, immunohistopathology, finite element modeling (FEM, and retrieval analysis. Screw holes allowed entry of titanium oxide and exit of polyethylene particles. Birefringent polyethylene wear particles were found behind the metal cup in macrophages containing pro-inflammatory tumor necrosis factor-α and interleukin-1β, whereas fibroblast-like cells stained for osteoclastogenic receptor activator of nuclear factor kappa B ligand (RANKL). Computerized tomography revealed granulomas (83% versus 17 %) and cortical destruction (50% versus 5%) better than radiographs. In FEM, a change of the abduction angle from 45 to 60 deg, and liner thickness from 4.8 mm to 2.5 mm, increased the back-side wear by 90% and 120%, respectively. Screw holes were stress concentration areas; their removal decreased wear by 40%. Modeling results were validated in retrieved implants, which demonstrated extensive back-side wear damage of liners with a high abduction angle. Combined clinico-radiological, immunohistopathological, FEM, and retrieval analysis disclosed that back-side wear in the HexLoc design is sensitive to the abduction angle, liner thickness, and presence of screw holes.