The implementation of the UK Control of Artificial Optical Radiation at Work Regulations2010 requires the employer to perform a risk assessment of workers’ exposure to UVradiation from phototherapy equipment in the hospital environment. The objective of thisstudy was to demonstrate that, where the source spectrum is known, an assessment ofexposure to UV sources commonly used in hospitals, including assessment of reflections andtransmission through personal protective equipment, may be performed with sufficientreliability by radiometer measurement rather than by complex spectroradiometricmeasurements. An intercomparison of radiometer and spectroradiometer measurementsof exposure to UV sources was carried out. Direct exposure was considered, aswell as exposure to reflected or scattered beams and that transmitted througheyewear. Assessment by radiometer of direct exposure and exposure to reflectionsdemonstrated an acceptable measurement error in the context of the inherent uncertaintyin the assumptions of the exposure scenario. Assessment of transmitted beamsmay result in a greater measurement error due to spectral mismatch; however,for typical exposure scenarios the error remained acceptable in comparison withthe exposure limit value. The methodology presented reduces the complexityof the measurement of UV hazard levels for common phototherapy equipment.