This work concerns neutron doses associated with the use of a Siemens Primus M5497 electron accelerator, which is operated in the photon mode at 15 MV. The conditions offer a situation within which a fraction of the bremsstrahlung emission energies exceed the photoneutron threshold. For different field sizes, an investigation has been made of neutron dose equivalent values at various measurement locations, including: (i) At the treatment table, at a source-surface distance of 100 cm; (ii) at the level of the floor directly adjacent to the treatment table; and (iii) in the control room and patient waiting area. The evaluated neutron dose equivalent was found to range from 0.0001 to 8.6 mSv/h, notably with the greatest value at the level of the floor directly adjacent to the treatment couch (8.6 mSv/h) exceeding the greatest value on the treatment table (5.5 mSv/h). Low values ranging from unobservable to between 0.0001 to 0.0002 mSv/h neutron dose were recorded around the control room and patient waiting area. For measurements on the floor, the study showed the dose equivalent to be greatest with the jaws closed. These data, most particularly concerning neutron distribution within the treatment room, are of great importance in making steps towards improving patient safety via the provision of protective measures.