BackgroundPET nuclides can have a considerable influence on the spatial resolution and image quality of PET/CT scans, which can influence diagnostics in oncology, for example. The individual impact of the positron energy of 18F, 68Ga, and 64Cu on spatial resolution and image quality was compared for PET/CT scans acquired using a clinical, digital scanner.MethodsA Jaszczak phantom and a NEMA PET body phantom were filled with 18F-FDG, 68Ga-HCl, or 64Cu-HCl, and PET/CT scans were performed on a Siemens Biograph Vision. Acquired images were analyzed regarding spatial resolution and image quality (recovery coefficients (RC), coefficient of variation within the background, contrast recovery coefficient (CRC), contrast–noise ratio (CNR), and relative count error in the lung insert). Data were compared between scans with different nuclides.ResultsWe found that image quality was comparable between 18F-FDG and 64Cu-HCl PET/CT measurements featuring similar maximal endpoint energies of the positrons. In comparison, RC, CRC, and CNR were degraded in 68Ga-HCl data despite similar count rates. In particular, the two smallest spheres of 10 mm and 13 mm diameter revealed lower RC, CRC, and CNR values. The spatial resolution was similar between 18F-FDG and 64Cu-HCl but up to 18% and 23% worse compared with PET/CT images of the NEMA PET body phantom filled with 68Ga-HCl.ConclusionsThe positron energy of the PET nuclide influences the spatial resolution and image quality of a digital PET/CT scan. The image quality and spatial resolution of 68Ga-HCl PET/CT images were worse than those of 18F-FDG or 64Cu-HCl despite similar count rates.